Cleaning Validation for Estimation of Vincamine Residues on the Pharmaceutical Equipment SurfacesImeda Rubashvili, Validation Department, Aversi-Rational Ltd, Tbilisi, Georgia
The method was developed using LC system “Ag 1260 Infinity” and Luna C18(2) 150 × 4.6 mm, 5 µm column with a mobile phase - a mixture of 0.1 M ammonium acetate solution and acetonitrile (40 : 60); The flow rate – 1.0 ml/min; The detector wavelength - 280 nm; The injection volume – 50 μL. The method was validated with respect to robustness, system suitability test, specificity, linearity-range, accuracy, precision (intra-day and inter day), limit of detection (LOD) and quantitation (LOQ). The stability of solutions of vincamine was also studied. These studies were performed in accordance with established guidelines. The calibration curve is linear r2=1.00000 over a concentration range 0.025 – 150 µg/mL; LOD - 0.005 µg/mL and LOQ - 0.025 µg/mL; The determined concentration of vincamine residues in swab sample solutions varies from 0.0069 - 11.60 µg/mL which is well below the limit of cross-contamination (123 µg/mL) of next product. |
|